| In the first place I would like to plead with our members to NEVER-EVER use a dedicated TDS (Total Dissolved Solids) meter, or the ppm scale on your EC meter. Always use the EC (or µS) scale.
The reason ? Well, there are no such things as tds-meters. Those devices that you stick in a watery solution, which then display a tds in ppm are in fact converted EC-meters, which measure the reverse electrical resistance of the solution, and the problem those meters have is that they cannot add ppm values together without serious errors. Example: Let’s say that you have a solution of 100 ppm Potassium Nitrate, and a solution of 100 ppm Calcium Nitrate. Suppose you mix the two solutions and stick in your tds meter to check, expecting to read 200 ppm, but . . . alas, because, depending on the meter you bought, it will either read more or it will read less, but never: 200 ppm (100+100). However, if you were to use the EC-scale you would read: 143 µS for the KNO3 solution, and 123 µS for the Ca(NO3)2 solution. And – surprise . . surprise – the mixed solution will read: 266 µS. Many members have come short on this confusing aspect of EC-meters, and many have adopted fallacious ways of suiting the ppm-scale. That wasn’t necessary. All that was needed is to change-over to the EC-scale. But how do you relate grams per liter (ppm) to conductivities ?
Example:
Note:
B) Tap water also has a conductivity, and you need to measure it before using, so that you can subtract its value from your reading to arrive at your nutrients concentration. You can also try to incorporate the calcium content into your formulas, but then, you need to know what you are doing.
|
Follow-Up Postings:
|
| Greystoke wrote:Example: Let’s say that you have a solution of 100 ppm Potassium Nitrate, and a solution of 100 ppm Calcium Nitrate. Suppose you mix the two solutions and stick in your tds meter to check, expecting to read 200 ppm, but . . . alas, because, depending on the meter you bought, it will either read more or it will read less, but never: 200 ppm (100+100). However, if you were to use the EC-scale you would read: 143 µS for the KNO3 solution, and 123 µS for the Ca(NO3)2 solution. And – surprise . . surprise – the mixed solution will read: 266 µS. This is not what I meant to say: The salts should be mixed together. Not the solutions. Doing it as descibed you would - of course - get the average of the two, ie: 133 µS, although the relative error would still be the same. Sorry about that. |
|
Just for those who don't want to get confused: The corrected version: In the first place I would like to plead with our members to NEVER-EVER use a dedicated TDS (Total Dissolved Solids) meter, or the ppm scale on your EC meter. Always use the EC (or µS) scale. |
Please Note: Only registered members are able to post messages to this forum. If you are a member, please log in. If you aren't yet a member, join now!
Return to the Hydroponics Forum
Instructions
- You must be a registered member and logged in to post messages on our forums.
- Posting is a two-step process. Once you have composed your message, you will be taken to the preview page. You will then have a chance to review the contents and make changes.
- After posting your message, you may need to refresh the forum page in order to see it.
- It is illegal to post copyrighted material without the owner's consent.
- HTML codes are allowed in the message field only.
- No advertising is allowed in any of the forums.
- If you would like to practice posting or uploading photos, please visit our Test forum.
- If you need assistance, please Contact Us and we will be happy to help.